In accordance with AWS D1.1, the test instrument, with cable and transducer is required to be recalibrated using a standard calibration block any time the cable or transducer is changed, any time a batter is changed, any time the instrument is unplugged and plugged into a different socket, and every 2 hours if I remember correctly.
I would expect ASTM E317 is similar.
Best regards - Al