One word of caution: Many years ago I worked on the neutral beam systems for one of the large fusion projects (DIII). We were pumping hydrogen, so depended on liquid helium cryopanels, but we also had large turbos backed by blowers and mechanical pumps with liquid nitrogen trap between pumps and chamber. When the beamlines were dismantled for upgrade to longer-pulse operation, my technician took some of the innards out for analysis, and they were soaked in oil. Turns out the oil from the mechancial pump worked it's way along the pipe walls through the pumps into the chamber. So, if you go to turbos, I would suggest you consider using oil-free mechanical pumps or, as alternative, a ten-foot or so PVC vertical pipe between turbo and blower -- oil doesn't adhere well to PVC, and won't be able to climb that high along the pipe walls.
Jim Treglio
Molecular Metallurgy, Inc.