MarkGallimore
Mechanical
- Oct 6, 2006
- 12
Hi, I've been asked to send a piece of laboratory equipment over to North America for use in a laboratory. It's intended to sit on a bench and plug into a receptacle. In the UK this apparatus has a 240v single phase plug which contains a 13amp fuse and all runs well - (note that the fuse protects the equipment and cable downsteam of the plug and not the upstream branch circuit). I understand that to power it in the US it will run off a 230v receptacle in the lab and will not be wired directly to a branch circuit as you might do with a large motor. The US plug will be added by the lab owner. I'm confused about fusing. In the current configuration the fuse is in the UK plug. If this is swapped for a North American plug then the only protective fuse will be that on the brach circuit (I think that's correct - could be wrong though). If the lab owner's branch circuit is fused with a 20 or 30 amp mcb then how does this protect my equipment in the event of a short or overload inside the equipment? Should I add a lower rated fuse inside my equipment or can the applicable US plug / receptacle hold fuses rated at a lower and more appropriate current for my equipment? In case it's necessary info the equipment is built to iec61010-1 and contains a heater and two servo motors and their controllers. Thanks for any advice offered.