Rocketmotor:
Regarding my reply of 22 Feb 08 please note the following:
The reason it has taken me so long to reply is that unfortunately, I loaned my copy of the Ling Electronics Vibration Training Manual document to a colleague and I am in the process of getting it back. When I obtain it I will forward it to those who have requested it.
In the mean time, the random vibration derivation I mentioned in my earlier post is presented in a document by Ling Electronics.
Vibration Training Manual
Ling Electronics
LTV Ling Electronics Division
Ling-Temco-Vought, Inc.
1515 So. Manchester Avenue
Anaheim, California
In the manual, for random vibrations, a displacement formula (equation 22) is derived. It is assumed that the elctrodynamic shaker and power amplifier are properly matched.
Equation (22) on page 2-28 of the training manual was derived as follows:
D = 42.8 * SQRT (G/(f^3)low)
D = double-amplitude displacement in inches (3 sigma)
G = the acceleration density in (g^2)rms/cps
f low = the lower cutoff frequency in cps
Based on this derived equation, I have listed a few possible scenarios as examples.
EXAMPLE 1:
* Suppose that our electrodynamic shaker is limited to 0.5 in. pk-pk displacement.
* Suppose a customer desires a random vibration test with an acceleration density of 0.4 g^2/cps.
* What is the lowest cutoff frequency of the spectrum, i.e. what cutoff frequency must we use for a 24-dB-per-octave filter?
* Solving for f low from the stated equation (22) we get f low = 14.3 Hz.
* If the shaker were capable of say 1.0 in pk-pk displacement then the lower cutoff frequency would be, f low = 9.02 Hz.
EXAMPLE 2:
* Suppose that our electrodynamic shaker is limited to 1.0 in. pk-pk displacement.
* Also suppose that the customer desires a random vibration profile with a lower cutoff frequency of, f low = 5 Hz.
* What acceleration density can not be exceeded?
* Again solving the stated equation (22) but now for the maximum possible acceleration density, G = 0.068 g^2/Hz.
EXAMPLE 3:
* Suppose that our electrodynamic shaker is limited to 1.5 in. pk-pk displacement.
* Suppose that a customer desires to perform a random vibration profile of G = 0.24 g^2/Hz and f low = 5 Hz.
* Can the test be performed on our shaker?
* Solving for the double amplitude D, yields D = 1.875 in.
* THE REQUESTED TEST CAN NOT BE PERFORMED since the required double amplitude would be too great.
I hope this helps.
Regards,
Tim Kaiser
VW Kaiser Engineering, Inc.