You should also be aware that mercury will form an amalgam with the porous gold electrode present in some moisture measurement sensors.
The resulting amalgam isolates the sensor's metal-oxide layer from the moisture in the natural gas resulting in a non-responsive sensor. This is particularly dangerous as the sensor will still output a signal thereby appearing as if it is still operating properly.
If your natural gas contains mercury, I would suggest using a technology other than a metal-oxide sensor.