Yes, and the tip about using diode test to check power diodes isn't what I would do either.
The resistance measurement method on p.6 is also something that shouldn't be done. Doing this, you include DMM series resistance plus test lead resistance and contact resistance in your result. Not recommended when you measure .5 - 5 ohms. Error will be considerable.
The varistor testing with an ohm-meter in Megohm range is not recommended: "A good Surge Suppressor should have a very high resistance,(more than 100 Megohms in either direction)." See p.11.
This is pure non-sense. Varistors shall be tested at rated current - usually 1 mA - and with short pulses. If that is not possible, see if the varistor limits inductive kick-back as described earlier. It does not have to be a large set-up. Use a bench PSU and a coil of any kind (transformer primary works well), connect varistor parallel to coil, apply and break DC, measure resulting peak with scope. Simple, easy, safe.
It is possible that the Stamford book is good for some, but I see it as a source of confusion. You have to know what you are doing and Stamford does not encourage that. It just gives tips that are mostly difficult to follow or even downright wrong.
Feeling quite grumpy now, I cannot resist asking if negative polarity diodes have left-hand threads? P.7.
Also, the resistance numbers for diodes are worthless if you do not specify at what voltage and current they are taken. That is why DMMs have diode test ranges, where diode fwd drop is measured. That measurement makes a lot more sense. But I still wouldn't use that on power diodes.
I could go on forever pointing out things that confuse the unexperienced technician, but I will stop with this observation: Polarity is different on DMMs and analogue meters. So the pictures on p.7 will confuse those who use DMMs.
Gunnar Englund
--------------------------------------