Joe Tank,
Errors in thickness measurements are solely a result of the test piece acoustic velocity being different than the velocity of the sample used to calibrate the instrumentation. According to the NDT Handbook, Volume 7 longitudinal velocities of steel range from 5,830 to 5,900 m/sec. Austenitic stainless steel velocities range from 5,660 to 5,740 m/sec. If an average of both ranges are used; inspecting a stainless steel component based on a calibration performed on mild steel will result in readings approximately 2.9% thicker than the actual component thickness.
Another concern is velocity variations within thick (several inches)stainless steel forgings and castings. I have seen velocity variations approaching 10% when the center of the piece is compared to the extremity. Even calibrating the instrument on the same piece then measuring an area with a different velocity will result in inaccurate readings.
In our shop ultrasonic thickness gaging is a method of lst resort, only used when direct measurement is not possible.
JR97