Tek-Tips is the largest IT community on the Internet today!

Members share and learn making Tek-Tips Forums the best source of peer-reviewed technical information on the Internet!

  • Congratulations MintJulep on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Fume Hood Testing 1

Status
Not open for further replies.

thegg1guy

Mechanical
Joined
Jul 26, 2007
Messages
2
Location
US
Does anyone know of any companies that do fume hood testing in the Pennsylvania, Deleware, New Jersey tri-state area ?
 
try looking up testing services on sites such as thomasnet.com
 
i've used a company named Precision Air Technology based in Raleigh, NC. Make sure whatever testing agency you use, uses an IR gas analzyer with a 0-20,000 cc/min sampling rate (usually set to match the isokenetic tidal volume of the average worker 15,000 cc/min). Do not use any testing companies that use the Qualtek leak detector, as their results will not stand up should you ever go to court with a worker exposure issue.

Also test as hard as you can and fail/eliminate any fume hood that doesn't meet your spillage rate at the "abuse" condition (normally full sash open).
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor

Back
Top