what instrument tolerance I need if I want to measure a process up to one decimal point?
 
Replies continue below

Recommended for you

Seriously. Not being versed in metrology I took the following fragment from Wikipedia:
A common convention in science and engineering is to express accuracy and/or precision implicitly by means of significant figures. Here, when not explicitly stated, the margin of error is understood to be one-half the value of the last significant place. For instance, a recording of 843.6 m, or 843.0 m, or 800.0 m would imply a margin of 0.05 m (the last significant place is the tenths place), while a recording of 8436 m would imply a margin of error of 0.5 m (the last significant digits are the units).
========================================================
Thus, accuracy can be said to be the 'correctness' of a measurement, while precision could be identified as the ability to resolve smaller differences.

I'd say the instrument should be able to measure down to the second decimal point.
 
The 10:1 test accuracy ratio (TAR) was previously imposed by MIL-STD-45662, but the current thinking per MIL-HDBK-52 is more like 4:1, so 1/4 of the minimum desired resolution is required.