Tek-Tips is the largest IT community on the Internet today!

Members share and learn making Tek-Tips Forums the best source of peer-reviewed technical information on the Internet!

  • Congratulations cowski on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Vacuum Gauge for 1 torr to 20 torr range: Manual Reading

Status
Not open for further replies.

plantprowler

Chemical
Joined
Aug 10, 2013
Messages
136
Location
IN
What are good options for vacuum measurement on a distillation column in the 5 to 20 torr range. The distillation column is small (10 inch dia; used for batch distillation of non-edible oils) & the measuring instrument is self standing. i.e. Not a part of a DCS / PLC / Control Loop etc.

Only a local operator needs to access the reading. Both analog / digital options are OK.

Accuracy requirements are very modest: A 2 torr accuracy is sufficient.

What options would people recommend? In the past the site has used mercury manometers which were working quite well. But now that the time for replacement has come I'm just re-evaluating all options.

PS. Are there aneroid / Bourdon type gauges that will work in the 5 to 20 torr range?
 
Most Aneroid and Bourdon gauges are differential gauges. They measure the differential between atmospheric pressure acting on the outside of the tube or chamber and the pressure to be measured.
I suspect that normal variations in atmospheric pressure may be more than the range of error that you will accept.
Take a look here;

Bill
--------------------
"Why not the best?"
Jimmy Carter
 
Status