Will clamp-on transit time ultrasonic flow meter work on your water line? Yes.
But the original post mentions the presumed effort involved with calibration to a master standard as driving the dissatisfaction with turbine meters.
Assuming the plant's requirement for calibraton is non-trivial, what constitutes a valid calibration? Certified traceable flow stand calibration?
Bucket and stop watch test?
What exactly does the "duplicate 'master' flow meter" accomplish in relationship to the working meters? Are the working meter sent out for flow stand calibration or somehow checked against a master turbine meter?
If the technology changes to ultrasonic, how would a valid calibration be accomplished?
The listing for clamp-on, the standard FUS1010 model transit time ultrasonic flow meter cal certs on the Siemens (formerly Controlatron) spec sheet (page 6) states:
- Wet flow transfer calibration, 6 point; based on nominal line size, up to 36"/DIN900.
Siemens has specific clamp-on ultrasonic flow meters designated as "Check Metering", described as "developed especially for verifying the accuracy and performance of any brand or type of flowmeter."
These models include a "certificate of intrinsic calibration". I'm not sure what that means.
Rather than doing an all around replacment of in-line meters, does the certification of clamp-on meters meets your requirement as check meter device? Could you consider using a single clamp-on meter to check the installed working turbine meters, rather than doing whatever it is that is done currently with the master turbine meter?