Thanks for your input.
I believe I have a firm theoretical grasp of the instrument. I have been testing on an FHWA project which requires moisture and density counts not only for the standard count, but for each individial test. I have observed that as the compaction increaces, these counts decrease.
One could draw the conclusion that the denser the fill, the less radiation is able to penetrate, and the higher your pcf. I will disregard the concept of false high readings away for the time being, because I believe those tend to be attributed to adverse chemical content of the fill- say if your soil contained H, as is the case with bottom ash from a coal plant.
Practically speaking, then, the only way I can see to get a false low reading is if the radiation was REFLECTED by a nearby structure, such as a pipe or concrete wall. Theoretically, this would add radiation sensed by the gauge that would otherwise be blocked by increaced density of the fill, and register as low compaction.
Though I respect and appreciate your opinions, in the field, I simply have not found this to be the case. A contractor will say, "Well, we're not getting compaction because of the wall".
I will say, "Well, why don't you try a second pass?"
The compaction will inevitably increace. I will leave with a passing compaction report, and the contractor, though he is probably counting the man hour that he lost, will probably appreciate the nice, flat road five years from now when he drives over it.
In short, I have simply seen too many inverted speed bumps on the road to give a contractor more reason to bend the rules.
Thanks for your comments, guys. I appreciate it.