Your Base Discharge is too high, if you have the Fanuc manual adjustment procedure is in there. Our Manual is B-70125EN/03 Section 9. Basically after you setup the parameters you adjust VR11 with 0 power commanded until discharge stops then back off until it just blinks on. You may trip breakers if you are too aggressive with adjustments. We just decommissioned a C2000 so if you ever need matched RF and DC units, hit me up, I think I could get owner to part with them much cheaper than fanuc or PLCCenter. i.fix.stuff at live. I don't make commision, just tired of the laser just taking up space.