Thermo vacuumpump error

Status
Not open for further replies.
Replies continue below

Recommended for you

check for all three phases at the circuit breaker/thermal relay for the vacuum pump. Disclaimer- don't electrocute yourself. reset the breaker/thermal relay

Chris Krug Maximum Up-time, Minimum BS
 
it's circuit breaker Q1 in the top back of the laser cabinet. Make sure you have all three phases. The turbo and everything else will run on single phase in some conditions.

Chris Krug Maximum Up-time, Minimum BS
 