As a displacement type sensor the tuning fork resonant frequency changes with the density of the fluid which will also contribute to a change in the shear rate.
Is it significant? who knows, having recognised that there is not a constant shear rate nor even one that can be quantified in any meaningful way, it would have been pointless to investigate it further, especially as it does not preclude it from being used as a process instrument for non-Newtonian fluids.
The shear behaviour is consistent enough to give repeatable results.
This is partly because (for the 7827) the time period is very short compared to the calculation period so any cyclic effects average out.
Because the frequency is comparatively high the data cycle is anywhere from 1-2 seconds (not to be confused with the calculation cycle which is user adjustable from 0.5seconds and up.
For example the device flips between operating at the upper and lower 3dB frequencies and the frequency of each is determined by measuring the time required to count 1000 cycles (typical). So a calculation period depends on the time for two such cycles (viscosity is a function of the bandwidth or the difference in the upper and lower 3dB frequencies) plus the stabilisation time (the time taken for the sensor to stabilise at the operating frequency). Hence any variation in viscosity due to variation in shear rate caused by the sinusoidal motion of the tines is not evident in the final measurement and the viscosity measurement can be very stable indeed.
The only surprise with some non-Newtonian fluids is just what that "viscosity" will be.
With any process instrument, even rotational viscometers where because of the different compromises in design between a laboratory and a process instrument, getting agreement between the lab and process instruments isn't straightforward, what is required is to find out what the process instrument says and then adjust its calibration to match the laboratory measurement (just like any other instrument?).
Two main sources of shear are the sensor shear and flow shear. Hence keeping flow constant provides a for a relatively constant shear regime for the measurement. It just won't be the same shear that you get with the lab instrument. It will be somewhere else along the viscosity shear curve and you correct for that in the instrument calibration software.
JMW