will123123

Materials
Nov 26, 2009
46
I need to get values of the specific heat capacity from a couple of gearbox oil samples at a range of temperatures (100-160°C). Initial measurements have been made using a modulated DSC, but the scatter in the results is excessive - this method is known to suffer from errors of at least 10%. Can anyone suggest a method with much improved accuracy.

Thank you
 
Replies continue below

Recommended for you

Standard calorimetry (i.e. bulk calorimetry) may be more what you are looking for? Although, DSC is supposedly much more accurate than 10%. Is it possible the scatter is due to real physical differences in the samples, e.g. due to breakdown of the oil?
 
At 160C vaporization of the sample is likely to be a problem. Are you using a pressurized cell?
 
Yes - sealed pans are being used during the tests. The results look good during a single test, ie. a consistent rise in SHC with temperature and there is no anomaly at the higher temperatures. Therefore I believe it is unlikely there is breakdown of the oil. The problem is scatter from one test to the next and I do not believe this is due to sampling errors (ie incomplete mixing of the oil).
 
There are three ASTM standards for determination of specific heat, two of which use a form of DSC:

ASTM D2766 - 95(2009) Standard Test Method for Specific Heat of Liquids and Solids

ASTM E1269 - 11 Standard Test Method for Determining Specific Heat Capacity by Differential Scanning Calorimetry

ASTM E2716 - 09 Standard Test Method for Determining Specific Heat Capacity by Sinusoidal Modulated Temperature Differential Scanning Calorimetry
 
When we did DSC on volatile samples we used an Ar pressurized cell.
Repeatability was a fraction of 1%.

= = = = = = = = = = = = = = = = = = = =