Tek-Tips is the largest IT community on the Internet today!

Members share and learn making Tek-Tips Forums the best source of peer-reviewed technical information on the Internet!

  • Congratulations MintJulep on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Quantify Accuracy of Load Cell

Status
Not open for further replies.

kayro

Mechanical
Joined
Jan 13, 2017
Messages
1
Location
US
Can't convince myself I'm doing this right.

0 - 25lb load cell
10V excitation
Rated Output: 2mV/V

nonlinearity: .10%
nonrepeatability: .05%
hysteresis: .10%
zero balance: 1.0%

E* = sqrt(.002^2 + .001^2 + .002^2 + .02^2) = 0.02 mv/V accuracy
*those numbers represent error % * Rated Output in mV/V

if I have 25lb/20mV full output at 10V excitation,
then my accuracy in lb is (25lb/20mV)*.02 mV/V = +/- .25 lb accuracy.

Disregard my amp and DAQ for the time being, I am just trying to assess the sensor. Thank you for any feedback, I can never straighten this out in my head.
 
Specifically, the hysteresis and non-linearity are not random errors. Also not clear why you doubled the error values in the RSS sum.
 
0.1% * 2mV = 0.002 mV

Zero balance isn't necessarily random either, depending on how you're using it, and how repeatable that is.

TTFN (ta ta for now)
I can do absolutely anything. I'm an expert! faq731-376 forum1529 Entire Forum list
 
OP, can you please post a link to the datasheet?

xnuke
"Live and act within the limit of your knowledge and keep expanding it to the limit of your life." Ayn Rand, Atlas Shrugged.
Please see FAQ731-376 for tips on how to make the best use of Eng-Tips.
 
Ahh, I read that as percent error, thanks Irstuff. And yeah, zero balance...
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor

Back
Top