Quantify Accuracy of Load Cell

Status
Not open for further replies.

kayro

Mechanical
Jan 13, 2017
1
Can't convince myself I'm doing this right.

0 - 25lb load cell
10V excitation
Rated Output: 2mV/V

nonlinearity: .10%
nonrepeatability: .05%
hysteresis: .10%
zero balance: 1.0%

E* = sqrt(.002^2 + .001^2 + .002^2 + .02^2) = 0.02 mv/V accuracy
*those numbers represent error % * Rated Output in mV/V

if I have 25lb/20mV full output at 10V excitation,
then my accuracy in lb is (25lb/20mV)*.02 mV/V = +/- .25 lb accuracy.

Disregard my amp and DAQ for the time being, I am just trying to assess the sensor. Thank you for any feedback, I can never straighten this out in my head.
 
Replies continue below

Recommended for you

Specifically, the hysteresis and non-linearity are not random errors. Also not clear why you doubled the error values in the RSS sum.
 
OP, can you please post a link to the datasheet?

xnuke
"Live and act within the limit of your knowledge and keep expanding it to the limit of your life." Ayn Rand, Atlas Shrugged.
Please see FAQ731-376 for tips on how to make the best use of Eng-Tips.
 
Ahh, I read that as percent error, thanks Irstuff. And yeah, zero balance...
 