I'm having problems with my self-built measuring device.
The input stage looks like this:
1V ------3M6-----------n.inv. amplifier x1,6------- 1V
| |
1,125M 5M4
| |
GND GND
When I connect a x1/x10 probe I get with 1V/10V applied to the tip 1V +-3mV at the output of the amp.
x1/x10 probe:
<-------9M---------->
| |
|__--____|
When I connect a x10/x100 probe I get with 10V at the tip and x10-range 0,987V?
x10/x100 probe (only x10 range shown):
<----4,49M--------->
|
|
0,969M
|
GND
Why are the results so different?