Yes. However, to settle an argument requires reputable references. The best two sources of information are ASME, Section V, Article 23, Ultrasonic Standards, under which ASTM E2491-06 is contained titled, Standard Guide for Evaluating Performance Characteristics of Phased Array Ultrasonic Examination Instruments and Systems.
Excerpt from E2491-06
4. Summary of Guide
4.1 Phased-array instruments and systems have similar
individual components as are found in traditional ultrasonic
systems that are based on single channel or multiplexed
pulse-echo units. These include pulsers, receivers, probes
and interconnecting cables. The most significant difference
is that phased-array systems form the transmitted ultrasonic
pulse by constructive phase interference from the wavelets
formed off the individually pulsed elements of the phased array
probes.
4.2 Each phased-array probe consists of a series of
individually wired elements that are activated separately
using a programmable time delay pattern. Varying the number
of elements used and the delay time between the pulses
to each element allows control of the beam. Depending on
the probe design, it is possible to electronically vary the
angle (incident or skew), or the focal distance, or the beam
dimensions, or a combination of the three. In the receiving
mode, acoustic energy is received by the elements and the
signals undergo a summation process utilizing the same
type of time delay process as was used during transmission.