Thank you for your responses. Some more information:
This electromagnet is wired in parallel with another electromagnet, and I noticed that it seems to be operating at higher current, relative to the other magnet, than it used to (using 89% of the total current vs. 78% previously, or 14% higher). It is difficult to replace so I need to be sure it is broken (internally shorted?) first. It operates DC, 100% duty cycle at up to about 50 amps. It is 8G square wire, 306 turns, 4.8" long by 3.2" in diameter. It is encapsulated in boron nitride containing epoxy and surrounded in circulating cooling water. It is part of a ion thin films deposition source. I have calculated its wire length and resistance based on the manufacturers data; and the current is about 14% high compared to calculated. What test do you recommend? What is "megger test"? Thank you.