Tek-Tips is the largest IT community on the Internet today!

Members share and learn making Tek-Tips Forums the best source of peer-reviewed technical information on the Internet!

  • Congratulations TugboatEng on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Explanation for change in magnitude after random vib test

Status
Not open for further replies.

Astorite

Mechanical
Joined
Mar 15, 2012
Messages
8
Location
GB
Hi,

As a small company and being new to vibration testing, we are having difficulty in explaining to our customer the reason for a change in magnitude on a post random low level Sine test. The background is that we complete a functional test on the equipment, Pre LLS test (0.5g @ 20-2kHz), Random vib, post LLS test and then a post functional test. Both X & Y axes were fine with correlation on the pre/post vib plots and the post functional test was fine for all 3 axes. However, in the Z axis, the magnitude at resonance reduced by approx 50%. There was a minimal change (~2%) in resonance for the pre and post plots. The unit wasn't altered nor moved during the z axis testing.

The customer specification allows a deviation of 10% on frequency and 40% on magnitude. As we are outside this, it's a test failure.

As far as an explanation, the reduction in magnitude suggests some settling of the unit in this axis provided additional damping.

Any other thoughts /experiences?
 
The server does not like the ampersand in your file name

TTFN
faq731-376
7ofakss
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor

Back
Top