Hi logbook,
Are you using one of the low-cost Hall-effect transducers? I have seen this problem before:
Some of the range of current probes and transducers manufactured by LEM-HEME have excellent bandwidth specs, but are prone to dV/dt pickup from the conductor under test. High dV/dt shows up in the output as apparent transients in the current. I first encountered this effect when making measurements on a current-source inverter, and at initially I believed I was seeing very high dI/dt in the circuit in spite of the presence of the huge inductance inherent in the design of a CSI. This defied any kind of reasonable explanation, and was ultimately disproven using one of the excellent (and expensive!) Tektronix AM-503 current probe amplifers and a P6304 (perhaps) probe. The Tektronix probe is virtually immune to dV/dt pickup, and showed the Hall-effect probe's shortcomings very clearly. The link between dV/dt on the conductor under test and the transients in output of the Hall-effect current probe was clear when conductor voltage was displayed alongside the output from both types of current probe on a multi-channel 'scope.
The Hall-effect probes can be made to perform better in terms of dV/dt immunity by a couple of tricks: monitor current on the 'earthy' conductor if possible - whether it is a DC return to ground, or an AC neutral conductor. This reduces the problem at source by keeping dV/dt small. The other trick is to create an earthed electrostatic screen between the conductor and the probe. I use the self-adhesive copper foil sold for EMC purposes, with a drain wire down to chassis ground. The foil can be applied to either the conductor if it is insulated, or to the bore and end faces of the probe. Do not totally cover the probe in this foil, or you will create a shorted turn through the ferrite core of the probe. With high current, high frequency measurements this can get interesting, as a colleague found out: the probe casing melted!
Hope this helps anyone using these otherwise good low-cost devices.
As an aside, anyone who uses the Tektronix current probe amplifier in an environment where high external magnetic fields are present, such as an inverter with large air-cored inductors, a word of caution:
The AM503 amplifier used with the current probe has an earthed chassis, as has the 'scope. The two are interconnected by a 50 ohm coaxial cable. A classical ground loop is formed by the protective earthing conductors in the mains power leads and the wiring of the building. Sketch out the ground conductors and you will understand. The output of the AM-503 amplifier is very low, and there can be enough current induced in the ground loop which adds (or subtracts) from the real signal to give some wierd results on the 'scope. I used an isolation transformer for the amplifier and removed the power ground connection, thus breaking the loop. The problem went away. Make sure you re-instate the ground connection when you are finished, and confine this sort of thing to controlled test environments.
----------------------------------
If we learn from our mistakes,
I'm getting a great education!