CMM calibration 1

Status
Not open for further replies.

pbboyd

Industrial
May 26, 2004
5
hi
we just purchased a used DEA manual cmm (1992) it is in excellant shape. when I set it up and tried to measure known dimensions it seems to read 0.001" out. I can't seem to find any reference in the manual to calibrate the machine or to compensate for this error. Does any know what to do.
Thanks.
 
Replies continue below

Recommended for you

Hello, I work for a metrology lab, and operate a B&S Global FX CMM with PC_Demis software. You must first have your reference sphere calibrated. This may be part of your problem. For instance,if your r-sphere is .74993 and the info in your software states it as .75000 Then this error must be corrected.Refer to your manuals on this matter. I suggest you have a technician come in and verify all axes of the machine and send your r-sphere to a cal-lab and have it verified. I hope this helps!

 