thanks.
I must agree with your comment, there are no ways to get the real dew point as the mirror needs some content of condensate (70 mg.m-3 is a std. value) to react, also mirrors can show errors of 5-7 C (MCMI v ACMI see the NPL REPORT AS 3), GC-EOS packages in general could be very accurate but they suffer from poor composition characterization, when tuned they need a mirror as reference, however this could be the laboratory's intrument.
In our case the GC manufacturer is proposing a quite expensive EOS package (>10000 US) so there are not real economical reasons to support this EOS solution... instead We are considering to install a EOS software within DCS and fix inputs for C6+ with the result from laboratory analysis, the preliminary results are promising, we plan to evaluate this solution for one / two months, collect results and create a quite large base of data to discuss with the clients, that could be a good starting point...
A think to mention : the EOS package would offer the advantage to calculate in realtime many properties that at present we obtain from our process simulator, I am not sure about final accuracy (must test and compare different software solutions) but this could be very useful for us...
A final note, we have evaluating different software packages and comparing dew point / cricondentherms as calculated with Soave Redlich Kwong and Peng Robinson, results are quite similar (we use our process simulator as reference) despite very large differences in price, this really surpises me.