The micro hardness testers, such as Mic 10 , do not use the classical Vickers visual measurement of the indentation dimension- they rely instead on the change in ultrasonic resonance of a small steel bar that terminated in a diamond tip. As the tip indents into the solid, the resonant frequency changes proportional to the hardness, but also in relation to the solid's young's modulus. Therefore, ehnever the solid has a young's modulus other than that of the default value, you must use a lab calibrated sample of the same material to be ultimately samples.
Case in point- P91 welded to 316H with a inconel 625 filler. The default calibration might be OK for the P91, but not the other 2 solids.