Hi, chromatogram shows FID (detector) response over time. Each response corresponds to a certain amount of VOC, and that correspondence is defined when you calibrate. So, first calibrate using standards containing target VOCs. Calibration curve could be three to six points, depending on the accuracy needed. Also pick convenient calibration range depending on your type of work, yet don’t forget about FID and column limitations. FID has a good linearity but you may need to do some dilutions if you are using pure or high products.
Look at it this way, lets say we want to read voc1, range 100ppm-2000ppm. I'd get voc1 standard and do, for example, (100,200,800,1200,1600,2000)ppm after each one FID would give response (best taken as area of the pick), for example, (xvfx,xxgfxx,xxxhghxxx,xxghgdxx,xxxxdrxx)then yourself or software plot response=function (amount). Now you are ready to measure.