Dear Friends,

I want to know if the velocity is one of the criteria when selecting the vortex flowmeter size.
Especially when the fluid is flammable gas or liquid.
In flowmeters, the flowmeter size could be less than the line size. So with a smaller flowmeter size, the fluid velocity increase.
With higher velocity, the possibility of building up static charge is high. This is a major concern while handling flammable fluids.
I would like to know of any guideline or thumb rule for such cases.


Thanks in advance
Replies continue below

Recommended for you

RE: Vortex Flowmeter Selection- Fluid Velocity

Velocity is critical factor for a vortex flow meter because vortex technology is inherently a velocity meter.

Go to this thread
[img https://www.eng-tips.com/viewthread.cfm?qid=436351]
And scroll way way down to the graphic of the velocity graph and the table of minimum flow rate values for various pipe sizes (for water).

At flow rates at or below the table’s minimum flow rate values, the flow meter outputs a zero flow rate value (4.0mA). At the low flow end, the flowmeter is in low flow cut-off mode because the formation of vortices is too infrequent to provide a useable flow rate value. Any brand vortex meter will have this inherent characteristic.

The manufacturer’s sizing program will indicate the useable flow rate range. Because it is a velocity meter, sizing usually shows better performance with line sizes smaller than nominal.

As an alternative, Thermal Dispersion technology tends to provide good meaurement at low flow rates so it is not as demanding for reduced pipe size meter runs or lengthy upstream/downstream steaight runs as vortex technology is.

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members! Already a Member? Login



News


Close Box

Join Eng-Tips® Today!

Join your peers on the Internet's largest technical engineering professional community.
It's easy to join and it's free.

Here's Why Members Love Eng-Tips Forums:

Register now while it's still free!

Already a member? Close this window and log in.

Join Us             Close