Hi,

I am using an Anritsu/Wiltron 54111A with the accessory SWR autotester to test SWR in the AMPS-CELL band, the GPS band and the PCS band.

This is a production test arrangement, and I want to test an antenna that I know is of about VSWR = 1.6 or so.  The problem is the antenna is at the end of a long cable (about 16 feet, but low loss -- about 2 db or less loss).  I also have a microwave Teledyne relay (rated at VSWR < 1.2) in the line near the antenna.  I also have a bias tee in the line to power the active GPS antenna.    I am cal-ing through all of this stuff, the bias tee, the cable, and the relay.  

 This Scaler NA uses only a short-open cal arrangement.   I try to cal with my standards at the  end of this 16' cable, and accurate measurements seem to be hit or miss.  Often the measured VSWR is way worse than what I know the antenna to be.  Yes, when I look at the display of freq vs VSWR, I do see "waves" corresponding to some phase effect --- I know this means something, but I forget exactly what.

Here's the odd thing....  I also have an HP Vector Network Analyzer, and when I cal through all this stuff (cable, bias tee, relay) I get perfect measurements every time.  

What's the difference here?   Why is the Scaler analyzer not giving good measurements?  

What is the HP doing for me that the Anritsu is not??

Thanks very much.

RE: Scaler Network Analyzer question

Scalar analyzers can really only function to the limits of the directivity of the couplers used to compare forward versus reflected waves. I nearly killed my self trying all sorts of multiple cal standards, de-embedding techniques etc. with scalar measurement data develping a custom high power TR environmental tester. As I remember, the key weakness is how all the various vectors align for any possible combination. Never the less, for these test I dumped the high power VSWR parameter. The vector analyzers allow the complete characterization of the black box two-port between the DUT and the "perfect" measuring terminals, I think it just boils down to number of equations = number of unknowns.  It gets really fun with a multi-port(>2) system, but that is another head-ache.

RE: Scaler Network Analyzer question

Here is what I would do.  Very accurately measure the insertion loss of the cable, switch, and whatever else you have between the Scalar analyzer and the antenna.  Lets say for argument that the loss is 2.0 dB.

Cal you scalar analyzer right at the analyzer.

Hook up the cable, etc., and the antenna.  Make a VSWR measurement.

Lets now say that the antenna just squeeked by with a real VSWR of 1.6:1.  That is a reflection coeficient of 0.375 and a return loss of 8.5 dB.  

Your test set sent out a signal, the cable attenuated the signal by 2.0 db, the antenna reflected the signal with a reflection loss of 8.5 dB, and the cable again attenuated the reflected signal by 2.0 dB.  So, if your reading on the scalar analyzer shows that the return loss is 12.5 dB, you know that your antenna had a 8.5 dB return loss.

RE: Scaler Network Analyzer question

VSWRs and return loss numbers are not much use for thinking about this problem. They both need to be converted to reflection coefficients to do the maths. Assuming a unit applied signal, or normalising to unity, the reflection coefficient is the reflected signal.

With
p=(VSWR-1)/(VSWR+1)

I get

VSWR=1.6 gives p=0.231
VSWR=1.2 gives p=0.091

Now all we have calculated is the reflection coefficient magnitude. The phases of these two reflected signals could be the same, opposite, or anywhere in between. We might see a reflected signal of
0.231+0.091 =  0.322        -->> VSWR= 1.95
or
0.231-  0.091 =  0.140        -->> VSWR=  1.33

Then we have this wretched 2dB loss in the cable. Any loss in the cable is going to increase the return loss, in other words the reflected signal is reduced. This make the VSWR look better. Suppose the cable had a 10dB loss. The minimum return loss would then be 20dB, and the worst possible measured VSWR would be 1.2.

The point is that having losses in the system makes the VSWR look better and reduces the accuracy of the system. You also cannot calibrate out the effects of the other components in the system when you use a scalar analyser. It should be possible to add a tuner to the system to get the VSWR to be 1 at one spot frequency without the antenna in circuit. You would then get a less inaccurate reading of the antenna. You would need to calibrate the system with a known mismatch though (preferably something around 1.6 VSWR).

I would say the test system is doomed to miserable and frustrating failure unless you use a VNA and calibrate it at the end of the test system at the point where the antenna will be connected using open/short/Z0-match calibration pieces.

RE: Scaler Network Analyzer question

Thanks, I did the math wrong.  I did .6/1.6 instead of .6 over 2.6.  Return loss is still the way to do it though.

RE: Scaler Network Analyzer question

I have not worked much with the scalar analyzers. But I have been to MTTS, etc tradeshows that demonstrated attenuation analyzers. These, even up in the mmw's would provide clean insertion loss curves from 80 to 120dB down. This was something I could never acheive with the VNA's. At least in the microwave band, not to mention mmw's. So what is the trickwith these units?

RE: Scaler Network Analyzer question

Insertion loss is an easy thing to measure. All you need is a big signal source, a sensitive signal detector, and a calibrated attenuator. The signal source doesn’t need to be calibrated. The detector doesn’t need to be calibrated. You just adjust the calibrated attenuator to give the same reading as the unknown when switching between the unknown and the calibrated attenuator.

The scalar and vector analysers when running in single-port mode have to extract the reflected signal from the incident signal, and do so with great accuracy. This requires high dynamic range and high directivity, both of which are not required for the insertion loss equipment.

RE: Scaler Network Analyzer question

(OP)

When I look at my plot of VSWR vs Freq, I see a periodic wave type of thing as I look along the freq scale.   In other words, the VSWR seems to differ with freq by as much as .8 or so VSWR.   I guess this is some phase effect as the electrical length of the cable changes with freq.   But I thought I would just be spinning around the smith chart without any effects in magnitude.

any thoughts??

RE: Scaler Network Analyzer question

You are (assuming a good measurement system) seeing reflections from at least two mismatches that are positioned at different physical points along the line.  Your analyzer can only read a composite reflected travelling wave, which is similar to a voltage.  Voltages add linearly.  So, if at some frequency the two major reflections arrive at your analyzer IN PHASE, there is a maximum voltage.  As some other frequency, these two reflections arrive at the analyzer 180 degrees out of phase, and one subtracts from the other, resulting in a voltage minimum.

RE: Scaler Network Analyzer question

(OP)
"seeing reflections from at least two mismatches that are positioned at different physical points along the line"


Very good, that makes total sense.   To extend my understanding of what's going on one step further, why is the VNA able to cal these effects out for a perfect measurement, while the scalar NA is not able to do this?

The VNA is an HP-8714B while the scalar NA is Anritsu/Wiltron 54111A.

Thanks!!!!

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members! Already a Member? Login


Resources

Low-Volume Rapid Injection Molding With 3D Printed Molds
Learn methods and guidelines for using stereolithography (SLA) 3D printed molds in the injection molding process to lower costs and lead time. Discover how this hybrid manufacturing process enables on-demand mold fabrication to quickly produce small batches of thermoplastic parts. Download Now
Design for Additive Manufacturing (DfAM)
Examine how the principles of DfAM upend many of the long-standing rules around manufacturability - allowing engineers and designers to place a part’s function at the center of their design considerations. Download Now
Taking Control of Engineering Documents
This ebook covers tips for creating and managing workflows, security best practices and protection of intellectual property, Cloud vs. on-premise software solutions, CAD file management, compliance, and more. Download Now

Close Box

Join Eng-Tips® Today!

Join your peers on the Internet's largest technical engineering professional community.
It's easy to join and it's free.

Here's Why Members Love Eng-Tips Forums:

Register now while it's still free!

Already a member? Close this window and log in.

Join Us             Close