Sterilization machine/cold spot failure

Status
Not open for further replies.

jumacao

Chemical
Joined
Dec 26, 2008
Messages
41
The range of the cold spot temperarure during an ETO (ethylene oxide) sterlization cycle is 45-55 C. The graph displays 44.9 C ones. What could be the possible causes for this parameters outside the expected range.
 
0.1 degree Celcius is only 0.2% of the value you are measuring. Does whatever instrument is measuring the temperature have better than 0.2% overall uncertainty? Is it calibrated? You are assuming the sterilising machine is out of spec, but unless you are using a precision instrument then you don't really know.


----------------------------------