Continue to Site

Eng-Tips is the largest engineering community on the Internet

Intelligent Work Forums for Engineering Professionals

  • Congratulations cowski on being selected by the Eng-Tips community for having the most helpful posts in the forums last week. Way to Go!

Sterilization machine/cold spot failure

Status
Not open for further replies.

jumacao

Chemical
Dec 26, 2008
41
The range of the cold spot temperarure during an ETO (ethylene oxide) sterlization cycle is 45-55 C. The graph displays 44.9 C ones. What could be the possible causes for this parameters outside the expected range.
 
Replies continue below

Recommended for you

0.1 degree Celcius is only 0.2% of the value you are measuring. Does whatever instrument is measuring the temperature have better than 0.2% overall uncertainty? Is it calibrated? You are assuming the sterilising machine is out of spec, but unless you are using a precision instrument then you don't really know.


----------------------------------
image.php

If we learn from our mistakes I'm getting a great education!
 
Status
Not open for further replies.

Part and Inventory Search

Sponsor