Orifice Meter - Causes of Bias

Status
Not open for further replies.

Zoobie

Chemical
Oct 22, 2002
193
I've got AGA3 in front of me and a bunch of other reference material but the answer I need is not jumping out at me. Could anyone out there give me the Readers Digest version of what factors contribute to bias errors as opposed to random +/- errors. Are there particular items that will consistently make the flow measurement higher than actual? The installations specifically are field gas gathering orifice meters.

Thanks in advance.
 
Replies continue below

Recommended for you

The only things that I've ever seen that shift field measurement high are incorrect parameter entry (e.g., wrong plate size, wrong pressure ranges), a shifted zero in a transmitter, or salt plugging in the plate bore. Everything else tends to shift it low. Some examples:
- backwards plate
- plugged sensing lines
- swirl in the flow
- bowed plate

I've attached a cartoon I use in my Primary Element talk to illustrate that the things we check in a calibration are just the tip of the iceberg.

David
 
 http://files.engineering.com/getfile.aspx?folder=8461c450-8ff4-4332-b6bb-0b8835b2678b&file=TubeInspectionCartoon.pdf
Thanks...and I really like the cartoon too.
 