I have a calculation I need to make based on a value of recorded pressure but need to know the error in the measurement. This could be measured either by Bourdon tube or by diaphragm/piezoresistive pressure sensor.
Typically for pressure sensors the manufacturer will publish an accuracy as a percentage of span and can be as low as 0.02%. A typical sensor may drift during its operational life and therefore require to be recalibrated against some means of test equipment within a given, traceable accuracy. In this application they are re-calibrated if during a 5-point check over the sensor span the measurement is out by over 1% of reading.
Assuming the sensor or gauge used has been re-calibrated before the measurement is taken against traceable test equipment how do you work out the accuracy of the sensor best case? (Worst case is assuming 1% of reading or worse).
Also, if the manufacturer publishes an accuracy of 0.02% of span is this only when first shipped or is it always 0.02% accurate?
Is the accuracy of Bourdon tubes any different?
Typically for pressure sensors the manufacturer will publish an accuracy as a percentage of span and can be as low as 0.02%. A typical sensor may drift during its operational life and therefore require to be recalibrated against some means of test equipment within a given, traceable accuracy. In this application they are re-calibrated if during a 5-point check over the sensor span the measurement is out by over 1% of reading.
Assuming the sensor or gauge used has been re-calibrated before the measurement is taken against traceable test equipment how do you work out the accuracy of the sensor best case? (Worst case is assuming 1% of reading or worse).
Also, if the manufacturer publishes an accuracy of 0.02% of span is this only when first shipped or is it always 0.02% accurate?
Is the accuracy of Bourdon tubes any different?