I was just as excited as you. Then I read the specs.
No more than 15 waveform memories - although they can store four traces each. But still no more than 15 separate measurements. The rest (also limited) needs to be stored as bitmaps. I must ask myself, again: WHY on Earth economize on something as important as storage space? An SD card would have solved that problem.
The math I described above can't be performed. Not possible to use more than +, - and x and only one result channel. I usually need three to make Three-phase measurements.
Accuracy! Have a look at the specs:
Vertical accuracy
5 mV/div to 100 V/div ±(2.1% + 0.04 range/div)
2 mV/div ±(2.9% + 0.08 range/div)
At 1V/div you get +/-(21 mV + 40 mV) = +/-61 mV, which corresponds to around 6% error margin. I need better. Doesn't matter if you go to 100 V/div - same error margin. I wonder if this is a mistake. It can't be that bad.
But, perhaps it can. The automatic measurements are shown with 999 maximum Count. That is not impressive. A tru 3 digit display - not even a "3.5" digit display.
AC accuracy (results shown on the 999 display is:
AC coupled, high frequencies
60 Hz to 20 kHz ±(2.5 % + 15 counts)
20 kHz to 1 MHz ±(5 % + 20 counts)
1 MHz to 25 MHz ±(10 % + 20 counts)
So, if you need to measure a 1 V 1 MHz sine, you can't trust the result to be better than +/-(1V*10% + 20) = +/-120 mV or +/-12%. Again, I must ask myself if there is a mistake in the specs.
Vertical resolution isn't good either. Just 200 pixels. The screen is 320x240 pixels.
I was, as I said, also in need of one of these. But I cooled off quite a bit after reading through the data sheets and the manual. I don't even think that a super low price tag could lure me into buying one. Sorry. We do need better than that. I expect the Chinese to bring a decent and easily affordable scope on the market. In the meantime, my "French mistress" works well for me.