I'm working on the safety analysis for CE Marking and this is the first time I've had a system that wasn't using a wall-wart for a power supply. So now I have to deal with the MAINS (note I put the 61010 terms in upper case) inside my enclosure going to the power supply.
My power inlet has a 6.3A fuse so it can handle my typical 1.8A load (5.3A if the supplies are 100% loaded). I have wires with BASIC INSULATION going to the power supply (hot, neutral, ground).
Under NORMAL CONDITIONS the basic insulation is fine. If I take a SINGLE FAULT CONDITION of the insulation breaking then the only protection is the fuse. IEC61010 Figure 1 shows that the fuse needs to blow within about 0.040 seconds, but that doesn't work with my typical load.
Am I missing something here? Luckily all of the MAINS wiring is inside the enclosure, so as long as I tell the user to stay out of the box nothing is ACCESSIBLE. That seems excessive though. Is the BASIC INSULATION failing the wrong SINGLE FAULT CONDITION? In reality it shouldn't be a problem unless someone pinches a wire under a bolt while assembling the unit. But that never happens...
My power inlet has a 6.3A fuse so it can handle my typical 1.8A load (5.3A if the supplies are 100% loaded). I have wires with BASIC INSULATION going to the power supply (hot, neutral, ground).
Under NORMAL CONDITIONS the basic insulation is fine. If I take a SINGLE FAULT CONDITION of the insulation breaking then the only protection is the fuse. IEC61010 Figure 1 shows that the fuse needs to blow within about 0.040 seconds, but that doesn't work with my typical load.
Am I missing something here? Luckily all of the MAINS wiring is inside the enclosure, so as long as I tell the user to stay out of the box nothing is ACCESSIBLE. That seems excessive though. Is the BASIC INSULATION failing the wrong SINGLE FAULT CONDITION? In reality it shouldn't be a problem unless someone pinches a wire under a bolt while assembling the unit. But that never happens...