Establishing a ground connection on a portable resistance test unit

Status
Not open for further replies.

switchguy

Electrical
Joined
Oct 17, 2003
Messages
7
Location
US
I recently purchased a Megger DLRO200 low resistance ohmmeter. Unfortunately the unit must be plugged into a 120 VAC source which is not convenient in our manufacturing facility. We purchased a pure sine wave inverter (300 W), battery charger and 12 V battery in hopes of converting to a self contained unit we could roll around on a cart. The problem we are facing is that the DRLO performs a test to verify a ground connection. The output on the inverter has a ground prong but aparrently not adequate for the DRLO to pass the test. I'm looking for any suggestions on how to get the DRLO into thinking it has a ground connection.

Thanks in advance,

 
Are you looking for a 'portable ground' ?!?

 
Low resistances? [neutral]

----------------------------------
image.php
Sometimes I only open my mouth to swap feet...
 
After getting to the right person at Megger I found out that I would have to establish a ground through the auxiliary ground terminals of the unit. A bit on an inconvenience but it will work.The ground connection is simply a precaution to assure that the circuit being tested is safe to test.

Thanks to all for their posts
 