Would welcome any advice / ideas as to possible cause to problem
Problem: During routine empty chamber requal of the autoclave, we are seeing a 4 degree C overshoot at start of exposure exceeding the HTM2010 limit of 3 degrees. This overshoot happened during heat up in previous requals ie it appears to have shifted by up to 30 secs. Also seeing an increased delta temp between the installed drain probes and chamber probes of 0.8C versus 0.3C during previous quals and an increased delta T of 3.08C from 1.5C across the validation probes. Also measure 6C superheat in header before the autoclave control valve per HTM2010 method versus 0 - 0.7C before.
Checked all possible caused (we could think of):
1 Measurement - recalibrated all probes to same working standard and confirmed several times - ok. We originally had a problem with the I/O card and repleced same and the CPU card was replaced because the battery voltage was slightly low - all control parameters checked to be ok. eproms not touched.
2 checked status of bleed valves and condensate bypass valve settings - ok
3 Jacket pressure switch dropped from 1.3 Bar to 0.8 Bar to eliminate Jacket superheating the chamber. Temporary change as is likely to lead to wet loads.
4 Control Valve operation - OK
5 PID values - altered the settings to get a compliant cycle but it is not the root cause and would likely lead to problems with distribution studies if left.
6 CSG functioning correctly - removed some lagging after PCV at CSG which is ~ 100m from Autoclave, to eliminate superheat & to see if it had impact on cycle - no significant impact - still failed cycle.
7 Cannot understand superheat at autoclave on CGS side of PCV as there is nothing to generate it and the temp corresponds to sat stm press in header. We use the HTM2010 method of measuring for superheat. Reconfirmed 3 times. No measurable superheat in chamber using packed towel test.
8 Need to find root cause - would really welcome any thought / suggestions
Problem: During routine empty chamber requal of the autoclave, we are seeing a 4 degree C overshoot at start of exposure exceeding the HTM2010 limit of 3 degrees. This overshoot happened during heat up in previous requals ie it appears to have shifted by up to 30 secs. Also seeing an increased delta temp between the installed drain probes and chamber probes of 0.8C versus 0.3C during previous quals and an increased delta T of 3.08C from 1.5C across the validation probes. Also measure 6C superheat in header before the autoclave control valve per HTM2010 method versus 0 - 0.7C before.
Checked all possible caused (we could think of):
1 Measurement - recalibrated all probes to same working standard and confirmed several times - ok. We originally had a problem with the I/O card and repleced same and the CPU card was replaced because the battery voltage was slightly low - all control parameters checked to be ok. eproms not touched.
2 checked status of bleed valves and condensate bypass valve settings - ok
3 Jacket pressure switch dropped from 1.3 Bar to 0.8 Bar to eliminate Jacket superheating the chamber. Temporary change as is likely to lead to wet loads.
4 Control Valve operation - OK
5 PID values - altered the settings to get a compliant cycle but it is not the root cause and would likely lead to problems with distribution studies if left.
6 CSG functioning correctly - removed some lagging after PCV at CSG which is ~ 100m from Autoclave, to eliminate superheat & to see if it had impact on cycle - no significant impact - still failed cycle.
7 Cannot understand superheat at autoclave on CGS side of PCV as there is nothing to generate it and the temp corresponds to sat stm press in header. We use the HTM2010 method of measuring for superheat. Reconfirmed 3 times. No measurable superheat in chamber using packed towel test.
8 Need to find root cause - would really welcome any thought / suggestions