Oxygen Flow in tubing
Help Required
I have a flow problem I would appreciate some help with.
I have oxygen @ 12 psi, a 12" length of 0.125" Bore pvc tubing.
Question ???
 What Will this deliver in Liters/minute or cubic feet/min
regards
Fifie
RE: Oxygen Flow in tubing
assuming operating temperature is approx 25 deg C
6 l/min would give you a pressure drop of 0.44 psi
mechat
What I am trying to find out is will the 0.125 2 ID tube deliver 15 Ltrs a minute of oxygen @ 12PSI
It is for a medical device the oxygen enters into a face mask
Pressure loss is not an issue
Fifie
No! to deliver 15 l/min you would need to increase the pressure or the tube id.
Mechat
What size would you recommend
Just to clarrify
Oxygen Flow required 15 Ltrs per minute
Maximum pressure 12psi
regards
Fifie
I would recommend 0.25" i.d. tube.
mechcat
many thanks
could you give me the correct way to do this calculation
regards