Handheld Meter Recommendations
I'm looking for recommendations on a digital handheld meter providing conductivity, pH, and temperature for field use.  The meters we are using now are expensive to repair and are not waterproof.  Their exclusive distributor network also keeps the purchase prices high.  Any recommendations?
Dean
RE: Handheld Meter Recommendations
A good source is The HACH Company.  Contact info - Internet: www.hach.com  PH: USA - 1-800-227-4224, International - 970-669-3050.  If they don't have what you need, nobody does.
I have worked with pocket pH probes quite a bit (Sorry for the long reply.)  I have not worked with conductivity meters.  For pocket instruments, they are usually separate.
You need to decide 1.)How accurate you want this probe to be (temperature compensation on the pH reading?  Two digits after the decimal point?) 2.) How much time you are willing to spend in calibration and 3) If size matters.  That will define your options.  When I was shopping I was looking for a waterproof pocket probe with temperature compensation which could be easily be purchased by several districts, with easy enough instructions that I would not have to personally train every operator.  If your requirements are different, your choice of probe will also be different.
Hach doesn't have a pocket pH probe, but does have some larger, higher-end probes.  Their products are usually good quality and operator friendly.  If they ever do come up with a field probe, I will certainly be looking at it.
The USABluebook catalog has a variety of pocket pH probes and conductivity probes.  I have worked with the Oakton pH Testr 2+ (one digit after the decimal, compact, temperature compensated but no temperature reading capability) and the pH Testr 3+ (two digits after the decimal, compact, temperature compensation and temperature reading.)  We have had numerous field problems with the pH Testr 3+.  They take a long time to calibrate and they have a tricky cleaning regiment that is not described in the instructions (I called the supplier to get the cleaning procedure.  I’ll happily share it if you choose that probe.)  I suspect a lot of our operator field problems are due to not cleaning the reference junction properly.  In comparison, we have had fewer problems with the pH Testr 2+.  If you need the extra digit of accuracy, though, you have to go for the more expensive one with the longer calibration time.
I have also worked with some of the Hanna probes.   I found that the calibration procedure for the pHep 4 was a bit strange.   (The probe has a display that says it should be placed in a pH 4 buffer, but if you put it in the pH 10 buffer it would be okay – I would have gotten way too many field calls if I had chosen that one.)   It was also less accurate.  It would easily be off by 0.2 pH units right after calibration.  The good thing about it was that it used regular batteries not watch batteries.
I was considering the Oakton pH/Con 10 water proof meter or the Myron L Ultrameter 6P.  Has anyone tried these?
we have 2 myron L 6P meters in our lab (powerplant:  used for boiler chemistry, cooling tower chemistry, cooling tower blowdown/cleanup, and of course boiler makeup).  Neither myron agrees with the other on a sample (of course they calibrate beautifully, but they don't read the same on a sample).  The one reads consistent with a Hach pH probe we have, while the other constantly reads higher.  They both have been back to the factory and read correctly when returned, but of course they are back to their differences again (within <6 months) and yes the pH bulbs are kept wet with a potassium chloride solution.
The myron L 6P that does read correctly is a great little meter, but the other one... well lets just say it has a lot to be desired.
Just my 2 cents.
I've been in contact with a lot of MyronL's and generally they're fantastic.