Coriolis meter for gas and liquid onsite proving

Coriolis meter for gas and liquid onsite proving

(OP)
Hi,

in the frame of geting a local regulatory approval, we are asked to do the proving of coriolis meters onsite. Some of the meters are for gas and others are for liquid. The problem is that the manufacturer said that there's no way to do the proving onsite while the local legal metrology insist to do the onsite proving. Is there a way to do the proving of coriolis meter onsite ? How to do it exactly (master meter ? tank prover ?) knowing that we don't have any testbed,

Thanks in advance,

RE: Coriolis meter for gas and liquid onsite proving

Do you have space to install the master meter in series with the test meter?

RE: Coriolis meter for gas and liquid onsite proving

(OP)
In fact i'm facing two cases, there is some locations where we can install a master meters other locations not.

RE: Coriolis meter for gas and liquid onsite proving

Can you try the clamp-on flow meter?
I used it to verify the condensate flow before.

RE: Coriolis meter for gas and liquid onsite proving

mk3223,
No agency is going to accept a clamp on meter to "prove" a custody transfer meter.

Jazimed,
Ron Beatty and I did a presentation at Southwest Research a few years ago that talks about many of the issues. Bottom line is that of course the manufacturer is correct and you simply cannot "prove" a coriolis meter. You might be able to drop in a V-cone (they don't take much space, don't require conditioned flow or long runs of straight pipe, and they are a fantastic flow conditioner) in front of the coriolis meters and show the regulators that the two meters provide the same numbers (they probably will in the liquid service, the gas service might surprise you how bad the coriolis is). At the end of the evaluation you might consider just tossing the coriollis meters and leaving the V-cones.

David Simpson, PE
MuleShoe Engineering

In questions of science, the authority of a thousand is not worth the humble reasoning of a single individual. Galileo Galilei, Italian Physicist

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources


Close Box

Join Eng-Tips® Today!

Join your peers on the Internet's largest technical engineering professional community.
It's easy to join and it's free.

Here's Why Members Love Eng-Tips Forums:

Register now while it's still free!

Already a member? Close this window and log in.

Join Us             Close