How to change the K-Factor on a endress hauser flow meter (from the local display)
How to change the K-Factor on a endress hauser flow meter (from the local display)
(OP)
Hii,
May someone help me with a way to change a promass 80 mass flow meter K-Factor. I can reach the K-Factor value from the sensor data, but unable to change it. Any suggestions to do it from the transmitter it self. I know that there is a service level access code... can someone share it wit me.
Thanks,
May someone help me with a way to change a promass 80 mass flow meter K-Factor. I can reach the K-Factor value from the sensor data, but unable to change it. Any suggestions to do it from the transmitter it self. I know that there is a service level access code... can someone share it wit me.
Thanks,





RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
Reaction to change doesn't stop it
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
Emerson meters are having an option of a built-in multiplier (M-factor) in their transmitters enabling you to adjust it after calibration without affecting the K_factor.
Any ideas... or advice about changing the K-Factor & what are the parameters that i have to respect while changing it.
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
Reaction to change doesn't stop it
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
item 6 of the operating instruction page 33. implies it is fixed at the factory, but then references a mysterious service code.
Looks like E & H need to give that to you - it's not he same as your personal code apprently
Remember - More details = better answers
Also: If you get a response it's polite to respond to it.
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
Custody transfer with square edge orifice plates requires plate replacement, but all other meters, flow calibration is to document the actual meter factor, not change it.
The conversion of the meter factor to actual flow is done in the monitoring hardware (transmitter/display/DCS/PLC, etc).
RE: How to change the K-Factor on a endress hauser flow meter (from the local display)
Reaction to change doesn't stop it