regarding the calibration duration
regarding the calibration duration
(OP)
dear,
I calibrated a detector on 10th Feb 2014, with a enclosed tag that addressed the due time is 10th Feb 2015, But after calibration the detector was not used anymore, only stored in the initial box and kept an calibration status. from 6th May 2016, we start to used this detector in working, But now the third part inspector says that the detector has out of calibration because of the due time tag.
the question is, after calibration, the useless time could be counted in the calibration duration, in my opinion, because the detector kept in an calibration status after calibration, so the time from calibration to the initial using day cannot be counted in the calibration duration. the calibration duration start time must be counted from the initial using time.
I calibrated a detector on 10th Feb 2014, with a enclosed tag that addressed the due time is 10th Feb 2015, But after calibration the detector was not used anymore, only stored in the initial box and kept an calibration status. from 6th May 2016, we start to used this detector in working, But now the third part inspector says that the detector has out of calibration because of the due time tag.
the question is, after calibration, the useless time could be counted in the calibration duration, in my opinion, because the detector kept in an calibration status after calibration, so the time from calibration to the initial using day cannot be counted in the calibration duration. the calibration duration start time must be counted from the initial using time.





RE: regarding the calibration duration
TTFN (ta ta for now)
I can do absolutely anything. I'm an expert! https://www.youtube.com/watch?v=BKorP55Aqvg
FAQ731-376: Eng-Tips.com Forum Policies forum1529: Translation Assistance for Engineers
RE: regarding the calibration duration
RE: regarding the calibration duration
The TPI is perfectly correct - the dates listed (calibration date and expiry date)are the only things that matter.
Regards,
Shane
RE: regarding the calibration duration
the project which will using the detector is postpone because of all actual factor.
I know it very clearly, the calibration interval cannot be recommended or determined by the calibration agency, the user of the detector shall keep the equipment in a good condition and in a valid calibration duration. but the calibration interval shall be determined by the user.
the small tag attached by the calibration agency which indicated an due time is invalid and cannot be regarded as the evidence of re-calibration.
if the qualify control program addresses the calibration interval, the user will re-calibrate the equipment as per the provision, if not, the user can determine whether the equipment need to be re-calibrated by himself.
this point is addressed very explicitly in ISO 17025
5.10.4.4 A calibration certificate (or calibration label) shall not contain any recommendation on the
calibration interval except where this has been agreed with the customer. This requirement may be
superseded by legal regulations.
RE: regarding the calibration duration
RE: regarding the calibration duration