Vacuum Gauge for 1 torr to 20 torr range: Manual Reading
What are good options for vacuum measurement on a distillation column in the 5 to 20 torr range. The distillation column is small (10 inch dia; used for batch distillation of non-edible oils) & the measuring instrument is self standing. i.e. Not a part of a DCS / PLC / Control Loop etc.
Only a local operator needs to access the reading. Both analog / digital options are OK.
Accuracy requirements are very modest: A 2 torr accuracy is sufficient.
What options would people recommend? In the past the site has used mercury manometers which were working quite well. But now that the time for replacement has come I'm just re-evaluating all options.
PS. Are there aneroid / Bourdon type gauges that will work in the 5 to 20 torr range?
RE: Vacuum Gauge for 1 torr to 20 torr range: Manual Reading
Most Aneroid and Bourdon gauges are differential gauges. They measure the differential between atmospheric pressure acting on the outside of the tube or chamber and the pressure to be measured.
I suspect that normal variations in atmospheric pressure may be more than the range of error that you will accept.
Take a look here;
http://www.omega.com/subsection/digital-pressure-g...
Bill
"Why not the best?"
Jimmy Carter