LP- Gas Flow Meter Calibration
Hello!
I want to ascertain if I can calibrate an LP- Gas liquid controls flow meter with water as medium.
RE: LP- Gas Flow Meter Calibration
David Simpson, PE
MuleShoe Engineering
In questions of science, the authority of a thousand is not worth the humble reasoning of a single individual. Galileo Galilei, Italian Physicist
"Gas liquid controls flow meter " does not make sense.
I suspect that calibrating a gas meter with a liquid flow is a wholly bad idea, but it is difficult to tell. This would only get you part of the answer, i.e. the actual volume going through it, but the real issue is how the gas meter resolves the pressure and temperature issues to get to a standard figure.
How low pressure is Low pressure?
If it is LPG, are you metering in a liquid phase, or a lower pressure gas phase, or both in any phase?
Here is the URL; http://www.lcmeter.com/en/Products/productPages/PD...
It gives you all the Tech data you need.
Thank you once again.
It is a Butane Liquid phase Positive Differential meter.
Chapter 4 - Proving Systems
Chapter 5 - Metering
Chapter 6 - Metering Assemblies
Chapter 11 Section 2.3 - Water Calibration of
Volumetric Provers
Chapter 12 Section 2 - Calculation of Petroleum Quantities
These chapters of the API Manual of Petroleum Measurement Standards supersede the API Standard 1101. Provide a means of conveniently diverting liquid for calibration purposes.
What is the purpose for the calibration? If this is for a custody meter, you want to have an expert calibrate it. There are firms that specialize in calibration.
Since the topic is calibrating a gas service meter with water, it might be of interest that Emerson contends that there is a case for calibrating gas service Coriolis meters with water:
White Paper: Applicability of water calibrations for gas service
partial snip:
Testing by numerous European and North American flow labs has confirmed the transferability of water calibration data on Micro Motion Coriolis meters to gas applications. Most notably testing sponsored by the Gas Research Institute in 2004 and documented
in report GRI-04/172 (see Figure 2), which covers water to gas transferability and wet gas performance of Coriolis meters. Conclusions in the report state, “The single fluid calibration tests show that a water calibration of a Coriolis mass flow meter can be used for natural gas applications without loss of accuracy”.
Industry testing has shown there is minimal benefit, from a calibration uncertainty perspective, in performing a gas calibration over a water calibration on Coriolis meters intended for gas measurement. Although this is the case, the user should review industry recommended
practices, standards, and regulatory requirements when establishing calibration policy for Coriolis.
Link to 6 page white paper:
Link
Also, this posting for advice doesnt say if this meter is in custody or fiscal quality metering service or just operational quality metering service - does this make a difference ?
