Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
I need someone to calculate Uncertainty in arriving at an output for these meters. These Heat Flow meters need to be accurate within 3% or so.

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
No I am not offering a job.
There have to be some Heat Engineers who do this Calculation for checking the Accuracy of Heat Flow Meters.
IT is given at Annexure 2 of ASTM C518 under EQ. 2.2.
Thanks

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

Somebody has definitely misunderstood something.
Either the OP hasn't understood that a discussion group like this is for people who help others to accomplish something for themselves, or...
Or member IRStuff didn't understand that you need to get this work done quickly and he'd better hurry up with the answer for you.

Well, instead of arguing about that, perhaps a more direct question is in order:

rkvbobby123, why aren't you doing the calculation yourself if you know what needs to be done?

STF

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

"Or member IRStuff didn't understand that you need to get this work done quickly and he'd better hurry up with the answer for you."

not without a written contract winky smile


see: http://www.lasercomp.com/Tech%20Papes/Papers/Error...

TTFN
FAQ731-376: Eng-Tips.com Forum Policies

Need help writing a question or understanding a reply? forum1529: Translation Assistance for Engineers

Of course I can. I can do anything. I can do absolutely anything. I'm an expert!

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
I don't know what is there to misunderstand. It was just an innocent query if someone has done those calculations and looked ar this annexure 2.2 under ASTM C518.
If not, not an issue. We just learn from each other. Basically, these Cals relate to finding R-Value of any insulation and I was looking at other folks work in this area.
The interacting Factors, S, L,E and Delta T have usual meanings given in the standard and equation is well spelled out there.
Thanks anyway.

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

I did, in fact, download the paper last night. The url is ill-formed Link

As for the misunderstanding, I quote, "I need someone to calculate Uncertainty," not, "I need help calculating" or "Has anyone calculated..."

TTFN
FAQ731-376: Eng-Tips.com Forum Policies

Need help writing a question or understanding a reply? forum1529: Translation Assistance for Engineers

Of course I can. I can do anything. I can do absolutely anything. I'm an expert!

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
I have this with me, infact lots of folks know him well.
Looking below then, could you interpret this table. I understand those percentages but how do we arrive there. We are comparing different materials I can see.
Thanks.

TABLE II. SINGLE-THICKNESS TESTS VS. SINGLE-THICKNESS
CALIBRATIONS
Tests Pyroceram® Pyrex® Vespel® Perspex®
6.35 6.60 6.18 5.00 mm
UP 63031 31618 14123 9505 V
LP -61795 -31012 -13797 -9242 V
Ref.values Calibrations
(W/mK) 6.35 mm X 2.091 0.875 0.476 W/mK
63031 V X 2.092 0.872 0.472 W/mK
-61795 V X 2.092 0.873 0.474 W/mK
4.011 Pyroceram® X 91.2 134.1 150.8 %
6.45 mm 2.15 X 0.468 0.255 W/mK
31618 V 2.15 X 0.466 0.253 W/mK
-31012 V 2.15 X 0.467 0.254 W/mK

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

TABLE II. SINGLE-THICKNESS TESTS VS. SINGLE-THICKNESS
CALIBRATIONS
Tests
Pyroceram® Pyrex® Vespel® Perspex®
6.35       6.60   6.18    5.00    mm
UP 63031   31618  14123   9505   V
LP -61795 -31012 -13797  -9242   V
Ref.values Calibrations
(W/mK) 6.35 mm X 2.091 0.875 0.476 W/mK
63031  V X 2.092 0.872 0.472 W/mK
-61795 V X 2.092 0.873 0.474 W/mK
4.011 Pyroceram® X 91.2 134.1 150.8 %
6.45  mm   2.15 X 0.468 0.255 W/mK
31618  V 2.15 X 0.466 0.253 W/mK
-31012 V 2.15 X 0.467 0.254 W/mK 

I'm going to try to re-present that data using a fixed space font, maybe it will make more sense. But, frankly, maybe not.

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
I tried to separate the numbers but it shows up like this again.
But Briefly, we are comparing four types.
My question was having calibrated with a atandard SRM, how does the mV reading play into the Final R-Value.
I know how to calculate Calibration Factor. How does it come in the picture.
63000 Micro Volts and W/mk and R-Value are related.
Thanks.
Have a great Thanksgiving.

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

No, uV has to do with the thermocouple reading. It's only relevant to the extent that it was what they used to make the temperature measurement. Had they used a PRT, would uV still be meaningful?

TTFN
FAQ731-376: Eng-Tips.com Forum Policies

Need help writing a question or understanding a reply? forum1529: Translation Assistance for Engineers

Of course I can. I can do anything. I can do absolutely anything. I'm an expert!

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

(OP)
The Calibration factor for a NIST SRM say is 1.2, atleast that was the case in my test.
The Same material of equal thickness should come close to this number.
If the material is of a different make than that of the SRM or of a different thickness, should the Calibration factor change or stay the same. What is the signidicance of Calibration factor apart from assuring the HFM is performing as it should.

I would appreciate if some Hand written cals are shown. I am of the Old School.

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

My two cents:

Uncertainty of any measurements is the RMS of the systematic and random uncertainties. say Utotal = RMS (Us^2 + Ur^2)

Systematic uncertainty is the sum of the all systematic uncertainties of the measurement including amongst other things the calibration uncertainty.
The calibration uncertainty is related to the correlation coefficient R2 of your linear regression or your linear fit (if it is linear).

Random uncertainty is basically linked to the number of measurements that at the end you average for a given sampling rate.

I dont know if that helps..

"If you want to acquire a knowledge or skill, read a book and practice the skill".

RE: Heat Flow Meters Laser Type: How to calculate Uncertainty in the Output R-Value

yep, thanks IRstuff

"If you want to acquire a knowledge or skill, read a book and practice the skill".

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources