Scaling Distance Error (Nikon Total Station)

Scaling Distance Error (Nikon Total Station)

(OP)
Hi everyone! I'm hoping someone can shed some light on an issue I'm having, because I've run out of ideas.

I'm a surveyor for a construction company and I've just moved to a new city and I'm attempting to get my current job control points all squared away. Here's my saga:

-First day here I checked the legal survey with my the Nikon 632 that they had on site. I came up with a shortfall of 8mm between two of their points (over 52m) and slightly more between two others that were 61m apart.

-I got the legal guys to come back and re-confirm their survey, which they did. Their Leica had been calibrated a week prior to coming to site so everything on their end should be perfect.

-I assumed the EDM on my machine must be out, so I sent it away for calibration and got a loaner Nikon Nivo.

-Same results.

-Got another loaner, same results.

-I moved the station inside, and on a flat floor, taped out about 45m. Station reported a distance that was 7mm short. I then tested at half that distance and the error was 4mm.

-I've tried these tests using 2 different backsight prisms, a brand new mini-prism, AND reflectorless.

-Temperature and pressure correction is on, prism constants are correct.

Both the rental company (Cansel) and a professional surveyor friend of mine I've talked to on the phone have no clue what the problem could be.

My only thought is that the stations I'm using are coming from 1 province over, which is about 500m about sea level higher than where I'm located. I don't imagine that would make much if any difference, but it's my only thought, at this point.

RE: Scaling Distance Error (Nikon Total Station)

(OP)
Your referring to the prism constant, correct?

If so, then yes. The prisms I'm using can either bet set to 0mm or 30mm, I've tried both. And the fact that the error is not constant (it grows with a longer distance) leads me to believe it's not prism related. I also tried prismless (with the same result), to further rule that out.

RE: Scaling Distance Error (Nikon Total Station)

It looks not impossible that some atmospheric correction could be unaccounted for: http://emtoolbox.nist.gov/Wavelength/Documentation..., but unlikely. Your few data points in your OP nearly fall on a straight line for error vs. distance, so there does appear to be a scaling error. You might want to check a fixed distance over time, recording temp, press, and RH, and see if there is a correlation to the error. One other possibility is to measure the distance concurrently with the "legal" guys to see if you can spot some sort of difference between their approach and yours. Is your area known to have anything usual in the air, like excess CO2 or whatever? Achieving parity down to 100 ppm is not necessarily trivial. What is the generally acceptable error for your control points?

TTFN
FAQ731-376: Eng-Tips.com Forum Policies

Need help writing a question or understanding a reply? forum1529: Translation Assistance for Engineers

RE: Scaling Distance Error (Nikon Total Station)

(OP)
Thanks for the feedback, IRstuff!

I accidentally found another guy in our district that also has surveying experience, and solid control points on his job site. I'll be taking our station out there and shooting his points on Monday so we can directly compare our machines and settings.

I actually doubt that any air density issues could really be the cause. I ran through the calculations on Friday and at best it would be a mm or two. For our control points I'd definitely not expect more than 3mm. I believe the error on the nikon is 2mm + 15ppm in precise mode or something similar. In the previous city/job I was at we never had a problem agreeing third party surveyors within 1mm. Coincidentally they used a Leica, just like the legal surveyors in this case.

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources


Close Box

Join Eng-Tips® Today!

Join your peers on the Internet's largest technical engineering professional community.
It's easy to join and it's free.

Here's Why Members Love Eng-Tips Forums:

Register now while it's still free!

Already a member? Close this window and log in.

Join Us             Close