Hi,

Can anyone explain the required level of density required along the complete radography and difference allowed between two specific locations. Will there be any sensitivity variation levels along with it??

Please reply

Regards
MNT

RE: Radiography Density and Sensitivity

Review Article 2 of ASME V.

RE: Radiography Density and Sensitivity

mnthiraviam,

Density and sensitivity are two different things. Density is a measure of the darkness of a radiograph. Sensitivity is the measure of the smallest detail which can be detected. Sensitivity is gaged by the penetrameter. That said, if the density of a radiograph is extremely light or dark it is more difficult (or impossible) to see the required penetramter thus affecting the measure of sensitivity.

Different specifications have different requirements for density. A common acceptable range of densities is 1.8 to 4.0. The density (or darkness) of the radiograph can vary throughout the radiograph depending on factors such as part thickness variation, part curvature and weld reinforcement. Some specifications have an additional requirement that the density of the radiograph at the location of the penetrameter must be close to the density in the area of interest. This requirement assures a potential variation in density would not affect the ability to see a defect based on the visibility of the penetrameter. Typical specification range for variation is the density of the area of interest can be +15% / -30% to the density at the penetrameter.

JR97

RE: Radiography Density and Sensitivity

(OP)
Thanks for the support

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources