Ultrasonic flow meter calibration
Ultrasonic flow meter calibration
(OP)
Hi,
I am installing ultrasonic's flow meters made by GE, I would like them to be caibrated before installation,
Any help with the calibration procedure please.
Thanks,
I am installing ultrasonic's flow meters made by GE, I would like them to be caibrated before installation,
Any help with the calibration procedure please.
Thanks,





RE: Ultrasonic flow meter calibration
You have not mentioned what is ur application. Whether its liquid/gas? You should mention details about ur application before you can get a satisfactory reply.
Whether u are buying inline or clamp on type USM from GE? Requirement of calibration would depend on lot many factors.
Regar
RE: Ultrasonic flow meter calibration
You are basically correct, but the only thing that sets off us old guys faster than missing information is text speak. "ur" for (I suppose) "you are" or "you're" is especially off putting.
David Simpson, PE
MuleShoe Engineering
"Belief" is the acceptance of an hypotheses in the absence of data.
"Prejudice" is having an opinion not supported by the preponderance of the data.
"Knowledge" is only found through the accumulation and analysis of data.
RE: Ultrasonic flow meter calibration
JMW
www.ViscoAnalyser.com
RE: Ultrasonic flow meter calibration
RE: Ultrasonic flow meter calibration
There are firms that calibrate flow meters. The firms pick up the meters, take them to test facilities, and certify the performance.
http://midamericameter.com/services
RE: Ultrasonic flow meter calibration
1) Locate 3rd party to calibrate flowmeter
2) Give them all your process information
3) Ship flowmeter to them
4) Wait for it to return
5) Review calibration certificate
6) Install calibrated flowmeter
It is a tried and true method.