Calomel Electrode - Negative PD

Calomel Electrode - Negative PD

Calomel Electrode - Negative PD

(OP)
After constructing a Calomel Reference Electrode, using; Hg + Hg2Cl2, few issues;

1. Negative PD (~ -0.2 V)

2. PD keep dropping and eventually drops to zero (within 24 hrs)

The standard reduction potential of calomel electrode is supposed to be +0.243V.

any thoughts?

RE: Calomel Electrode - Negative PD

I am trying to do the same thing. Are you sure you have your leads the right way? You would have +0.2V if they were swapped.

I was getting -0.500V until I noticed that my silver pin was touching the Hg/Hg2Cl2 mixture and messing up the redox reaction.

How are you making the Hg calomel "paste"? I have seen both a paste and a method with both being seperate. I am also making mine very small to fit in a medical instrument so packaging is an issue for me.

Fill me in with more details, maybe we can help eachother out.  

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources