Heating of the RTD wire leads
3
I'm working on a temporary fixture that will use an RTD to measure the temperature of a flat surface.  The design might allow for the wires to the RTD to heat up to about 300F.  Will this effect the performance/accuracy of the RTD significantly?
Any insight will be helpful.
RE: Heating of the RTD wire leads
I doubt it.
First, the allowable temperature  of the wiring  to the RTD probably limits the temperature of the RTD environment but heat transfer from the wires should be miniscule.
50 ft distance using 30 gauge copper adds 10.3 ohms
230°F rise changes that resistance by 5.3 ohms
5 ohm error in 100 ohm platinum RTD is about 2°C error
TTFN
Use a 3- or 4-wire RTD if you are concerned.
Thanks for your help however, I do have another questions.
The RTD is not fixed to the surface that it is measuring; contact will be done with slight pressure and no thermal conductive epoxy between the RTD and measured surface will be used. Can I expect accurate results?
Thanks for your help.
Thermocouple T (range -328 °F to 752 °F) could be an option. There are available models specifically conceived for flat surfaces.
most modern rtd's are flat unless they've neen encapsulated
how to you intend to account for radiation effects in your measurment?