Fluke 2176A thermocouple thermometer

Fluke 2176A thermocouple thermometer

Fluke 2176A thermocouple thermometer

(OP)
Hello.  I have a Fluke 2176A thermometer that is calibrated to type T thermocouples.  I have K thermocouples.  The unit is supposedly also good with K thermocouples, but I do not see how to let it know there are K thermocouples attached.  Does anyone know about this thermometer?  I thought I'd ask before tearing the thing apart to try to tell.  It seems if I put a thermocouple different than a T-type, I will get an inaccurate reading.  Right?

Thanks!

RE: Fluke 2176A thermocouple thermometer

The thermometer can be coupled both to type T and to type K thermocouples. I do not think you need to select the type of thermocouple, just use the same type of thermocouple for all channels available.
The discrepancies you've detected when turning from T-type to K-type are in the nature of the thermocouple itself. Thermocouples type K and T have a different range  of applicability. For type K (-210 °C to 1370 °C) while for type T (-210 °C to 400 °C). Their accuracy is different as well. The reference standard is IEC 584-2. If you have not access to this standard take a glance at the link below
http://www.microlink.co.uk/tctable.html
 

RE: Fluke 2176A thermocouple thermometer

Depends on what temperatures you try to measure.  At 150°C, a type K would read about about 12°C lower than a type T.  You can use a number of online calculators to experiment: http://www.hartscientific.com/funstuff/tc_its90_calc.htm

At 350°C, the reading error would be about 96°C

Closer to room temperature, the differences are less than the inherent uncertainties in the measurement itself.

TTFN

FAQ731-376: Eng-Tips.com Forum Policies

RE: Fluke 2176A thermocouple thermometer

(OP)
OK, I did a bit of searching myself.  The 2176A is internally set on a circuit card and it is not easy to change thermocouple types, it appears.  It definitely DOES make a difference on the type of thermocouple used since each type will supply a different voltage and if the machine thinks it is a T thermocouple, the temperature reading will reflect that.

This is my plan:  use the K-type thermocouples with the T-type thermometer.  The reading will be a temperature resulting from a supplied voltage due to the temperature the thermocouple "sees".  Using the cool converter supplied by IRstuff or charts that can be found online, like here:

http://www.omega.com/temperature/Z/zsection.asp

(See thermocouple reference data section)I can convert the read temperature to a voltage on the T-chart.  I then go to the K-chart and use the voltage to find what the real temperature is.

Let me know if you guys think I'm crazy and I'm missing something!

Thanks!

Red Flag This Post

Please let us know here why this post is inappropriate. Reasons such as off-topic, duplicates, flames, illegal, vulgar, or students posting their homework.

Red Flag Submitted

Thank you for helping keep Eng-Tips Forums free from inappropriate posts.
The Eng-Tips staff will check this out and take appropriate action.

Reply To This Thread

Posting in the Eng-Tips forums is a member-only feature.

Click Here to join Eng-Tips and talk with other members!


Resources